I have Fluke 5102B calibrator,which is working fine with its D.C ranges but it does not works with its A.C. ranges.The o/p is exactly 2.14 times higher than the programmed value.Readings are as follows:
Programmed values at 1kHz o/p @fluke 8060A DMM
100mV 0.2128V,967.7Hz
1V 2.135 V,967.7Hz
10V 21.40 V,967.6Hz
100V 213.9 V,967.7Hz
400V 867.2 V,967.5Hz
100microAmps 0.2136microAmp,967.7Hz
1mA 2.133mA
10mA 21.36mA
100mA 213.1mA
500mA 1070.0mA
It's Ohm ranges are showing stable & correct for 10,100,1k,10k Ohms,but for 100k,1M,10M does not stable.It varies around its expected values.
For 100kOhms o/p varies between 99.79 to 100.22kOhms same for 1M &10M
Manufacturing date of the calibrator is around 1987.
Can you guide me which component exactly I have to replace.
Thanks with regards
KIRAN
Hi Kiran,
I am sorry,I was away, unable to reply you and regarding problem with Fluke calibrator 5102 B, I am sorry,you have to contact Fluke and their contact Tel 01603256690, because we do not deal with any repair or calibration.
Thank you for your enquiry,
Regards,
Devinder Sohal,
Farnell technical,
Hi Jim,
I am trying to calculate a Mean Time Between Failures (MTBF) for a system using MIL HDBK 217F, unfortunately I still have a few components left which I can't seem to find the MTBF for, nor do I know how I should categorise them to use the MIL HDBK to get an MTBF for them. The components I am struggling with are:
1. Octal Buffer, Line Driver.
2. Amp Isolation
3. Buffer / Line Driver
4. Temp Sensor
5. Power Distribution Switch
6. CPCI Bus Hot Swap Controller
I have included links for the components below. If you could tell me what section these fall under in the MIL HDBK or anything else that would help me predict the MTBF you would be doing me a huge favour.
http://www.linear.com/product/LTC4244
Thank for your time,
Peter
HI Jim
Would you have any recommendations on a replacement for the Fluke 8920A true RMS RF voltmeter ?
My requirements are true RMS millivolt measurements centered around the 400kHz region.
The nearest I have found is the Rohde&Schwarz URE3 but this is costly, and the Fluke 85RF probe adaptor is not sensitive enough.
Any ideas suggestions ( that don't involve buying second hand 8920A's ) welcome.
Regards David Mordey.
Jim,
Can you tell me how to capture the milliseconds time to trip RCD Circuit Breakers? I believe you would need a trigger point to start the timer and store the time value once the shunt trip has opened the CB. Can this be done without a Power Oscilliscope?
I have a Fluke 9100 Calibrator but it is limited for RCD tests.
Basil
Jim,
I would like some help in trying to find the right firm to develop a light meter that currently is not manufactured by any vendor. We have some very specific requirements for a meter used in a niche business.
~Cody
© 2009 Premier Farnell plc. All Rights Reserved
Premier Farnell plc, registered in England and Wales (no 00876412), registered office: Farnell House, Forge Lane, Leeds LS12 2NE