E36313A Triple Output DC P.S. - Review

Scoring

Product Performed to Expectations: 8
Specifications were sufficient to design with: 10
Demo Software was of good quality: 10
Product was easy to use: 10
Support materials were available: 10
The price to performance ratio was good: 10
TotalScore: 58 / 60
  • RoadTest: E36313A Triple Output DC P.S.
  • Buy Now
  • Evaluation Type: Power Supplies
  • Was everything in the box required?: No - I received a unit which was configured for a 115V mains supply and it was supplied with a US power cord. Being in the UK I have a 230V mains supply and our plugs are very different. This is not a problem as there are switches on the back to configure the voltage and I have other power cables available.
  • Comparable Products/Other parts you considered: There are numerous power supplies available from all the main test equipment manufacturers. The most direct comparison product I have available is a TTi QL335T or alternatively I have a TTi PL303QMD + TTi CPX400S combo, both of these solutions would give comparable triple PSU outputs.
  • What were the biggest problems encountered?: The unit being shipped with the wrong configuration but this isn't really a problem as it's easy to solve. I have found that there is an audible mains hum from my unit regardless of whether the supply is turned on or off, for as long as the AC inlet is powered. From searching online I found that whilst most people report the unit to be extremely quiet, there are some people who have units which make a mains hum the same as my unit. I reported the issue to Randall and he contacted Keysight who have sent me an email this morning confirming that this is not normal for these units and they are going to send me a replacement unit and arrange for my current one to be collected so they can investigate the issue. This is excellent service from Element14/Keysight.

  • Detailed Review:

    1. Introduction

    1.1. Overview

     

    I received my shiny new Keysight E36313A Power Supply from Element14 this week. Thank you so much to both Element14 and Keysight for picking me for this Road Test. It's an excellent product and I am very happy indeed to have one sitting in my lab. I have a pretty good setup already but my current PSU's are missing a few really useful features which the Keysight unit happily provides.

     

    This Power Supply is the top of the range model from the Keysight E36300 range. There are three models currently available, the entry level E36311A which is limited to 80W total power output and has a limited feature set, and the fully featured E36312A and E36313A models which have all the features enabled but with 80W and 160W total output power respectively. These models can also optionally have a GPIB interface which is supplied in my review unit. Unfortunately I do not have any way to utilize GPIB currently, but as the unit also has USB and Ethernet this is no problem at all. The main differentiations between the available features for the three models is shown in the image* below:

     

     

    * Taken from the Keysight E36313A Bench Power Supply Data Sheet

     

    1.1.1. External Links

     

    Keysight E36300 Triple Output Bench Power Supply Data Sheet: http://literature.cdn.keysight.com/litweb/pdf/5992-2124EN.pdf

     

    1.2. What's in the Box?

     

    The unit came very well packed in a sturdy box and held firmly in place within the box with custom packaging. Inside the box was the E36313A itself and a small box containing the following additional items:

     

    • US Power cable - This was a mistake as I was shipped a unit intended and configured for the US market. More about this below.
    • A bag containing 1 x 4-way 3.81mm terminal plug, 1 x 4-way 5.08mm terminal block plug, and 1 x 8-way 5.08mm terminal plug. These are to connect to the rear panel I/O connectors. More about this below.
    • A bag containing 1 x Keysight Instrument Control DVD. This contains the IO Libraries 2017 Update 1, Command Expert 2017 Update 1, and the BenchVue 2017 suite.
    • An envelope containing the Certificate of Calibration for the unit.
    • Various additional pieces of documentation, e.g. Packing List, Quick Start, Safety Information, etc

     

    1.3. The Main Specifications

     

    The main specifications for the three versions available of the Keysight E36300 Bench Power Supplies shown in the image* below:

     

    * Taken from the Keysight E36313A Bench Power Supply Data Sheet

     

    1.4. Comparable Solutions

     

    For this road test I am going to be evaluating the Keysight E36313A and comparing it to two other solutions I have available to me:

     

    1. TTi QL355T - Like the Keysight, this is a triple output unit.
    2. TTi PL303QMD + TTi CPX400S combo - In order to have the same three outputs I could combine two other power supplies. In this case my dual output TTi PL303QMD and my single output TTi CPX400S units.

     

    SpecificationKeysight E36313A

    TTi QL355T

    TTi QL355TP

    TTi PL303QMD

    TTi PL303QMD-P

    TTi CPX400S

    TTi CPX400SP

    Price*

    * As on Farnell UK 18/12/2017

    £1317.00 + VAT

    £760.00 + VAT

    £947.00 + VAT

    £529.00 + VAT

    £707.00 + VAT

    £558.00 + VAT

    £717.00 + VAT

    DC Output Ratings

    Channel 1: 0 to 6V @ 0 to 10A

    Channel 2: 0 to 25V @ 0 to 2A

    Channel 3: 0 to 25V @ 0 to 2A

    Channel 2||3: 0 to 25V @ 0 to 4A

    Channel 2+3: 0 to 50V @ 0 to 2A

    Channel 1: 0 to 15V @ 0 to 5A / 0 to 35V @ 0 to 3A / 0 to 35V @ 500mA

    Channel 2: 0 to 15V @ 0 to 5A / 0 to 35V @ 0 to 3A / 0 to 35V @ 500mA

    Channel 3: 2.7V @ 0 to 1A / 3.3V @ 0 to 1A / 5V @ 0 to 1A

    Channel 1: 0 to 30V @ 0 to 3A

    Channel 2: 0 to 30V @ 0 to 3A

    Channel 1||2: 0 to 30V @ 0 to 6A

    0 to 60V @ 0 to 20A (Max 420W TPD)
    Max Total Power160W110W180W420W
    Operating Modes

    Independent, Tracking, Series*, Parallel*

    * Fully automatic connection and control

    Independant, Linked*, Tracking

    * Enables linked controls with manually configured series / parallel modes

    Independant, Tracking, Series*, Parallel*

    *Series with manual connection in Tracking mode,

    Parallel is configured internally with a switch on the front panel.

    Independant
    Load Regulation - Voltage< 0.01% + 4 mV

    Channel 1: < 0.01% + 2 mV

    Channel 2: < 0.01% + 2 mV

    Channel 3: < 1.0% for a 90% load change

    Channel 1:

    Channel 2:

    < 0.01%
    Load Regulation - Current< 0.01% + 500 uA

    Channel 1: < 0.01% + 250uA (5A/3A range), < 0.01% + 50uA (500mA range)

    Channel 2: < 0.01% + 250uA (5A/3A range), < 0.01% + 50uA (500mA range)

    Channel 3: < 1.0% for a 90% load change

    Channel 1:

    Channel 2:

    < 0.05%
    Line Regulation - Voltage< 0.01% + 1 mV

    Channel 1: < 0.01% + 2mV for a 10% line change

    Channel 2: < 0.01% + 2mV for a 10% line change

    Channel 3: < 0.1% for a 10% line change

    Channel 1:

    Channel 2:

    < 0.01% for a 10% line change
    Line Regulation - Current< 0.01% + 500 uA

    Channel 1: < 0.01% + 250uA (5A/3A range), < 0.01% + 50uA (500mA range)

    Channel 2: < 0.01% + 250uA (5A/3A range), < 0.01% + 50uA (500mA range)

    Channel 3: < 0.1% for a 10% line change

    Channel 1:

    Channel 2:

    < 0.01% for a 10% line change
    Output Ripple

    Channel 1: < 350 uVrms / 2 mVpp

    Channel 2: < 1 mVrms / 5 mVpp

    Channel 3: < 1 mVrms / 5 mVpp

    Channel 1: < 350 uVrms / 2 mVpp

    Channel 2: < 350 uVrms / 2 mVpp

    Channel 3: < 1 mVrms typical

    Channel 1:  < 400 uVrms / 2 mVpp

    Channel 2: < 400 uVrms / 2 mVpp

    < 3 mVrms / 15 mVpp
    EthernetYesNoYes (P version)Yes (P version)
    USBYesYes (P version)Yes (P version)Yes (P version)
    GPIBYesYes (P version)Yes (P version)Yes (P version)
    RS-232NoYes (P version)Yes (P version)Yes (P version)

     

    So we can see from the above whilst the solutions are comparable in some respects, there are differences which may make one solution or another a better fit for certain applications. For example, the combination of the TTi PL303QMD-P and the TTI CPX400SP give a massive total power capability of 600W between the two which for some higher power applications may be an advantage.

     

    However, it should be noted that it's not all about the fine detail of the specs, a huge factor for a general purpose lab power supply is the ease of use and its ability to be well integrated into an overall test solution. We can see from the above that making various configurations for the output power rails is much simpler with the Keysight as all the connections as well as the combining of controls is all handled internally to the power supply with no additional wiring necessary.

     

    3. First Look at the Unit

    3.1. Front Panel Overview

     

    Looking at the image to the left we can see the unit has a fairly simple layout. There is a large screen to the left with size buttons immediately below it in a row.

     

    To the right there is a section with two rotary knobs, one to control the voltage and one to control the current. Above these are three colour coded buttons to select which of the three output channels is selected for control. Immediately below these are some navigation and control buttons plus a numeric keypad.

     

    Across the bottom of the unit you can see the binding posts for all three outputs. These are also colour coded and you can see they match the colour coding of the buttons at the top and their individual output enable buttons. It should be noted that, although you can't see here, the coloured buttons illuminate when the unit is powered on.

     

    To the left of the screen there is a USB port to enable a flash drive to be connected.

     

    Last but not least at the bottom left hand side of the front panel we find the soft touch power button and it's associated power on LED.

     

    3.2. Rear Panel Overview

    From the image to the left we can see that the rear panel of the unit is quite well utilized, there certainly isn't much room to add anything else! A large area in the middle is taken up by a vent for a cooling fan. I've heard this fan spin up a little during startup but not much at other times. Maybe once I get into testing this unit under higher load the fan will kick in and become more noticeable but generally the fan itself is very quiet. I'll touch on the noise levels from the unit a little bit more below.

     

    At the top left there is an 8-way 5.08mm terminal block connector for connecting to Output Channel 2 and Output Channel 3 of the power supply. Note that this provides four pins per supply output.

     

    Working our way down, we find an earth stud and then down at the bottom another 4-Way 5.08mm terminal block connector, this time for connecting to Output Channel 1 of the power supply. Again, note that this provides 4 pins for the supply output.

     

    At the top right we have the LAN connector and the USB connector. This USB port would be used to connect to a computer to allow the instrument to be controlled over USB. In my case I will be using Ethernet for my control interface so this will not be required. It should be noted that there was NOT a USB or Ethernet cable supplied with the unit and I think for a unit in this price bracket these should definitely be supplied. Fortunately after a short rummage in a drawer I managed to dig out a spare Ethernet cable so my review did not come to a screeching halt at that point.

     

    At the bottom right you will see the power inlet. At this point I should mention that I am based in the UK. The eagle eyed amongst you may notice there is something amiss noted on the back of this unit. You can see that the 115V configuration marking is shown. This unit was actually shipped to me configured for the US market and as such was set up for 115V rather than the 230V we use in this country. Fortunately I saw this before plugging in and powering up the unit otherwise I expect I would be requiring a replacement.

    The unit is easily configurable to support different voltages. As you can see there is a table on the back showing the switch positions for 100V, 115V, and 230V. In my case I simply needed to slide the right hand switch over from the left to the right to change it to the correct setting. Below is a close up on this section of the rear panel to give a clearer view of the configuration switches. So, a small disaster has been averted and I can get on with reviewing my unit!

     

     

    4. Getting More Familiar with the Unit

    4.1. Teardown!

     

    In the words of Dave Jones from the EEVBlog (like you didn't know who he was?!), don't turn it on, take it apart!

     

    Okay so I'm not Dave Jones, or any of those that get given lots of free kit with the explicit expectation they'll tear it down. I'm not going to totally strip the unit down into as many pieces as I can get it into now because I don't want to risk accidentally damaging the unit and I really like this unit and want to be able to use it as my main lab power supply so I am not going to do anything to risk breaking it. If Keysight would like to send me a faulty unit they have lying around, I would be more than happy to do a more detailed tear down of that at a later date, but for now I'm just going to pop the lid and take a few pics of the insides. It's quite easy to get the lid off if you have a Torx driver set as it is held in place by two Torx bolts on each side and two on the top at the back. Once these are removed the lid simply slides out backwards.

     

    4.1.1 The Main Board

     

    Unfortunately I am being very cautious so this is the best view I can get of the main board without taking the unit apart further and looking at what, if anything is located on the other side. I can see some residue on the board so it's not been quite as well cleaned in production than I would have liked but it's probably no-clean flux so not going to be causing any issues.

     

    You an see a few features of this board, for example there are two separate diode bridge's taking the secondary AC and then converting to various DC levels. There appear to be regulators for +12V and -12V, +3V3, +2V5, some current monitoring and towards the bottom I can see circuitry relating to the constant current and constant voltage references.

    4.1.2. Front Panel Controller Board

     

    You can see inside here, it's a typical embedded processor board. The main processor which controls the unit is shown and we can see it's all running on an STM32. You can make out a few other pieces of circuitry, some local power supplies, clocks, the FPC ribbons which go to the LCD and a coin cell battery presumably for an RTC or for a hold-up on some memory being used to store settings. I'd go for it being for an RTC as these days there are lots of options for non-volatile storage for small data sets like settings which are really cost effective and don't require an external hold up supply.

    4.1.3. A Deep Dark View into the Back!

     

    There is not a lot else to say about this one, I tried to shine some light in to the back to try and illuminate what was there, but unfortunately there is just too much in the way. You can just about make out that there is another board at the bottom of the unit, running across the back.

     

     

     

     

     

     

    4.2. A Run Down of the User Interface

     

    We can see from the image on the left that the unit has a very clear full colour LCD. The firmware uses appropriate colour coding for each of the channels which ties in with the colour coding for the buttons and the terminal post surrounds. This all ties the various displays and the physical controls together nicely and makes it very easy to quickly see for which channel you are viewing or editing the settings.

     

    Along the bottom of the LCD there is an area reserved for button labels. This area is fixed and there is one segment for each of the physical buttons which reside immediately below the LCD. The labels help guide you through the various options and update to be specific for each page you are viewing.

     

    Editing the settings for a particular channel is a simple matter of pressing a button to get to the control page for the currently selected channel. From there you can use the navigation buttons to move to each of the various options, you can type values in on the numeric keypad, or you can use the voltage and current knobs to make fine adjustments quickly. For setting a specific voltage or current it is almost always easier just to dial it in on the keypad than to try to use the knobs.

     

    Across the bottom there are four buttons, three of these are colour coded channel output enable buttons as previously described and the fourth is a button dedicated to turning all the channels on or off at the same time.

    4.2. Delving into the Usability

     

    So the big question is, how well does this interface work for every day usage scenarios. Well, I've only had it a couple of days so haven't used it extensively at all yet but it seems very intuitive to use. The combination of a simplified single set of controls to operate all three channels, the colour coding and the very large and clear colour LCD give a good foundation for a easy to operate unit. Lets now take a short tour around a few of the menus to see how we'd go about setting up the unit to perform a few basic tasks. We'll do this by way of some examples which we will work through to show the various steps required.

     

    So before we get into the actual setups, I wanted to show you a image of the main screen and show a little glitch I found when using the web interface (we'll look at this later) to grab images from the screen.

     

     

    Looking at the two images above, these are both of the main LCD screen when it just has the default configuration of 0V and maximum current set up on all it's outputs.

     

    If you look carefully you will see that on the first image there is a glitch in the current reading for both channel 1 and channel 2. These are reading exceedingly small values, in the order of <10uA and there is naturally some noise on these readings. However, you'd expect the capture to grab a stable image. It would appear rather than the capture grabbing the currently stable display buffer it is grabbing a buffer that is the subject of the next display update. I don't know if the rendering to the display itself is double buffered or not, although I haven't seen any tearing on the display which is the usual sign of updates happening on the live display, but this issue with the screen capture is something to watch out for if you are grabbing images for a report. It means rather than working through quickly and grabbing all the images you need as you go, you have to switch to another window and check the grabbed image and recapture if necessary. I hope Keysight fix this issue soon!

     

    In the screen captures you can see the information about each channel which is displayed, i.e. The set voltage and current, the actual measured voltage and current, whether a 2-Wire or 4-Wire connection is used, and whether the channel is turned on or off. If I now press the Channel 1 select button again the display will change to give more detailed information for Channel 1 as shown below.

     

     

    We can see that the Channel 1 area has now expanded to take up 2/3 of the display area with the Channel 2 and Channel 3 displays being condensed to show only their measured voltage and current and stacked on top of each other.

     

    With this configuration the display shows a lot more information about the channel setup including over-current and over-voltage protection settings (OCP and OVP) as well as any delay settings relating to the OCP and also for turning the output On/Off.

     

    Finally we see it now also displays the measured power in W too which is a really nice feature which should speed up things like generating efficiency curves for DC-DC converters (especially if I had a Keysight Bench DMM with a similar feature!) or checking the power consumption of a circuit quickly in various operating states.

     

    Above I made a brief comment about showing whether 2-Wire or 4-Wire output modes are being used. In case you are not familiar with this terminology, I will explain further now.

     

    • 2-Wire simply means that only the + and - outputs of the PSU go directly to the load which means that depending on the load and the cables used there may be a significant voltage drop in the cables and the voltage measured at the supply output may not be what is presented to the load.
    • 4-Wire connections use an additional pair of wires as a "sense" connection. These carry no significant current so they measure the exact voltage at the load. This measurement is then used to adjust the powers supply output to compensate for the voltage drop in the cables to ensure the load sees the intended voltage. Please note: If you set 4-Wire and your sense wires aren't connected with some supplies it is possible to significantly over-voltage the load. I'll take a look at how this supply deals with this scenario in due course.

     

    Those with eagle eyes (or generally anybody who has got this far and been paying attention!) will notice that the front panel doesn't have the capability available for 4-Wire connections. If you wish to use this mode of operation then you will need to make some cables up with the supplied terminal plugs and use the rear panel connectors.

     

    4.2.1. Set Channel 1 to 5V with a 2.5A current limit.

     

     

    We're starting from a clean slate with this so currently all the channels are set up with the default configuration so we can see that Channel 1 is the selected channel and is currently set to 0 Volts @ 10 Amps. In order to configure the voltage and current for this channel we simply press the button below the 'Source Settings' label to get to the source configuration screen for Channel 1.

     

    Once in the 'Output 1 - Source Settings' screen we can see that there are a number of things that can be configured beyond just the basic voltage and current.

     

    We can configure the various protection thresholds for over voltage (OVP) and over current (OVP) and we can configure delays for the over current protection. You'll notice that for the over current there are two possible delays, 'OCP Delay Start' and 'OCP Delay'. The reason for a separate start delay is so that during power up, where there is likely to be a significant inrush current we can increase the delay before the OCP trips to allow the system to power up correctly, but then once it's up and running we can introduce a faster OCP trip if desired. For our example we aren't looking at OVP and OCP so we'll leave these settings at their default values.

     

    We can also choose between 2-Wire and 4-Wire connections in this screen. We'll be using the front panel connectors so we're going to leave the default 2-Wire setting.

     

    Using the Up/Down/Enter buttons and the numeric pad we can now go ahead and enter the values for the desired output voltage and current limit.

     

    Once everything has been set up the resulting screen should be as shown to the left.

     

    Once we return to the main screen we can see that our new settings for Channel 1 are shown in the 'Set' boxes at the bottom of the Channel 1 display area.

     

    You might have noticed that Channel 2 is now selected. This is because I selected this channel to move onto the next set of tests before I remembered to save the screenshot. It's not some random mode change or anything odd like that, it's just me almost forgetting to grab what I needed for this report.

     

    4.2.2. Set Channel 2 and 3 to provide +9V and -9V rails which are tied together and tracking each other with a 1A current limit.

     

    The same procedure as we used for Channel 1 above will be used for setting the desired voltage and current for channel 2. It's all very quick and intuitive to get set up and it would be very difficulte to get it wrong. If you try and set a voltage or current which is out of range it will tell you and there is visual feedback at every step at to what changes have been made.

     

    Here I have configured Channel 2 to be 9V with a current limit of 1A.

     

    Returning to the main screen we can see Channel 2 has the correct settings, but at the moment Channel 3 is not set up.

     

    At this point we need to press the button below the 'Output Settings' label to take us to the configuration screen which will allow us to configure the modes available.

     

    On this screen we can configure various options, On/Off delays, On/Off coupling, Inhibits and also the mode of operation.

     

    For these examples we will ignore the delays, coupling, and inhibits and we'll come back to these later. For now we will look purely at the 'Operation Mode' option which allows us to select between four different modes as described in the next section.

     

    If we press the button below the 'Operation Mode' label we will move to the screen below.

     

    Here we can see several pieces of information:

     

    1. Down the right hand side is a summary of the current output settings.
    2. A large picture and description is shown in the main section which gives a picture of what is available at the various outputs along with information about exactly what is available from the various connections shown.

     

    Along the bottom the button labels have changed to show the various possible modes:

     

    • Independent - The outputs are completely independent and can be set to any valid value for each output.
    • Series - The outputs are internally connected in series with the output voltage being across the positive pin of Channel 2 and the negative pin of  Channel 3.
    • Parallel - The outputs are internally connected in parallel with the output voltage being available on the Channel 2 output pins.
    • Tracking - Channel 3 is set to track the settings of Channel 2 so anything that happens to that channel will be reflected on Channel 3.

     

    We want the outputs to track each other because we want a +/-9V set of power rails. In this case we've pressed the button below the 'Mode Tracking' label.

     

    The picture and description below show us what is now available from the Channel 2 and Channel 3 outputs.

     

    Pressing the 'Back' button we return to the main screen and can see that Channel 3 has now changed to show 9V as it is tracking Channel 2.

     

    One thing to be wary of is that only the voltage setting is reflected when choosing tracking. I think it would be useful to be able to set the current limit based on the tracked channel too, this could avoid some annoying mistakes. It's always good to check these things before applying power, fortunately the way this information is presented with this unit makes it really easy and quick to spot these kinds of things.

     

    We can correct the output current limit for Channel 3 by setting it in the same way we did for Channel 1 and Channel 2.

     

    To the left we can see the result of setting the current on the 'Source Setting' screen for Channel 3.

     

    Finally on the main page we can see that out outputs have now been correctly configured.

     

    In order to achieve the +/-9V which we desire on the output we shall externally connect the positive of Channel 3 to the negative of Channel 2 and use this also as the 0V reference for our circuit. Channel 2 will now be our +9V supply and Channel 3 will now be our -9V supply.

     

    4.2.3. Set Channel 2 and 3 to produce a single 48V rail with a 0.5A current limit.

     

    In this setup we are going to be configuring the power supply to put Channel 2 and Channel 3 is series mode. This allows double the maximum output voltage of any one of these two channels. Of course we could do this with most decent bench power supplies with floating output channels simply by tying the positive of one channel to the negative of the other.

     

    The beauty of the Keysight E36313A though is that connecting the outputs in series or parallel mode is all done internally so it gets configured correctly and doesn't rely on any external wiring which could be a point of failure (or mistake) in a test set up. Also in these modes, the main controls of channel 2 directly control the output voltage so no need to mess about setting multiple voltages and current limits. Simply set it how you want it as if it's a single power supply channel with those specs.

     

    So when I did these setups, my head was in the usual mindset for setting up series outputs manually. In this case I set Channel 2 to be 24V at 0.5A first, intending to have to go and do this with Channel 3 also but once I had done this and selected 'Mode Series' below, I soon realised this wasn't necessary!

     

    In the output setting window we can how choose 'Mode Series'. It will update the page to show how to connect to the supply in this instance and the limits of the combined channels. In this case we effectively have a single output channel which can supply up to 50V at a current of 2A which equates to 100W maximum.

     

    Going back to the main page it clearly shows that the two outputs are connected together and shows the combined voltage and current setting for these channels as a whole. Very nice!

     

    4.2.4. Set Channel 2 and 3 to produce a single 24V rail with a 4A current limit.

     

     

    Much like we did above for connecting the two output channels in series, we can select parallel mode. In this mode the single resulting output channel is able to provide a maximum of 25V with a current of up to 4A which again equates to 100W maximum.

     

    If I press the button to choose 'Mode Parallel' you can see the display has updated to show how we now connect to the supply. In this case we connect across the Channel 2 outputs and leave the Channel 3 outputs unconnected as these supplies are connected together in parallel internally.

     

    Going back to the main screen we can see it has updated to show we are now in parallel mode.

     

    We can see that we have the 24V @ 0.5A originally set on Channel 2 but paralleled up with the same setting on Channel 3 to give us a total of 1A on the combines Channel 2 output.

     

    If we now go into the usual screen we can see that we can now set the current to 4A which is the combined maximum current of the two channels in parallel.

     

    Finally back to the main screen and we can see the resulting setting with the Channel 2 and Channel 3 outputs combined into a single output and a clear indication it is in parallel mode.

     

    4.3. The Web Interface

     

    The image to the left shows the main view of the web interface on the unit once you have browsed to the unit on the network and logged in. As you can see it is an almost exact clone of the main layout of the front panel of the physical unit. The buttons on the unit correspond exactly to clickable buttons on the web interface so the way you operate the web interface and the way you operate the physical unit are almost identical. The only slight difference is there are buttons labelled with arrows for turning the control knobs up and down rather than having a gesturing method to allow the knobs to be operated more naturally. This would be a benefit if using the web interface on a tablet as it would enable a more natural feel when using a touchscreen. But this is a minor niggle really in an otherwise very clear user interface.

     

    I would also like the option to be able to make the unit view full screen rather than it being only a part of a larger page. Again, if trying to use this interface from a tablet this would be a big advantage as currently it'll be a little more fiddly to use on the smaller screen with so much real estate taken up by non-critical information.

     

    At the bottom of the screen below the unit to the left is a screen capture button. The purpose of this is to grab and save a BMP image of the current screen of the unit, i.e. it's a grab of the contents of the LCD screen area only, not an image of the whole front of the unit. I really like this idea and will help with documenting tests much more easily. Unfortunately there is a small niggle with this which is that the screen capture doesn't appear to use any double buffering so it will quite happily grab a screen capture mid update so if you have a reading which is not stable you can get a capture with a mix of numbers before/after update which is quite frustrating. This should have been spotted and fixed early on, preferable before the unit was released as it's quite an obvious bug. None the less, it's still an excellent feature which I will use a lot, even if I do have to grab a couple of extra shots each time to get one that's not got any refresh related image errors.

     

    The other thing I would like to be able to do from this interface is specify a directory to download the captured screenshots as currently it just puts them into the browsers default download location. As well as this I would like to see them employing an image numbering system. With the current set up when I press the screen capture button the first time I get a file called screenshot.bmp. If I take another capture it creates a file called screenshot (1).bmp, then screenshot (2).bmp and so on. Maybe a better way would be to make the capture filename a combination of the unit's serial number and the current data/time such that even with multiple units available there would never be a conflict in the file names generated.

    4.4. BenchVue

     

    I had intended to take a look at how this supply integrates into KeySights BenchVue instrument control and automation suite. However, upon launching my Windows 10 virtual machine where I had previously installed an used the software I found it had disappeared which is totally weird. I then downloaded the installer for the latest BenchVue 2017 and tried to reinstall it and finally after about 4 hours it appeared to complete the installation, however upon looking for anything relating to BenchVue being installed I found almost nothing except for the BenchVue FieldFox app, none of the other apps and not the main application. Again, this was totally weird and disappointing. Fortunately I was able to resurrect a Windows 7 virtual machine from elsewhere and I found on this that BenchVue 2017 installed quite happily and comparatively quickly. Happy days!

     

    Unfortunately the happy feeling only lasted a short while. After loading the latest version of BenchVue I found that the majority of the modules for it are now only available at cost with the addition of Pro licenses. This is not a change I particularly like as the earlier version of BenchVue I had gave access to feature limited versions of the various modules so you could at least get a feel for its operation. I have this power supply, my oscilloscope and my multimeter, all of which are separate modules so the cost to buy these is significant. There is a trial version available for each of these but as I don't know how I might wish to use this yet I don't want to start the trial until such time as I am ready to invest some serious time into investigating and setting up BenchVue.

     

    4.4.1. External Links

     

    BenchVue Software 2017 Technical Overview: http://literature.cdn.keysight.com/litweb/pdf/5991-3850EN.pdf

    BenchVue Software 2017 Platform Website: https://www.keysight.com/main/software.jspx?ckey=2417463

     

    5. Lets Do Some Experiments!

    5.1. Powering-up the unit, how fast does it start?

     

    The first thing I noticed when I powered up the unit was that even though it's a smart power supply with a processor running a network stack and providing a web interface and various programmable options, as well as supporting a modern user interface with a large LCD, it powers up and is ready to use within just a few seconds. Below is a video showing the overall boot up speed of the unit.

     

     

    So as you can see, from me starting the camera recording, pressing the button to switch on the power supply to me stopping the recording, it's only 10 seconds in total, the power supply booting up is only a few of those seconds which is very nice. Nice one Keysight!

     

    5.2. Testing the Voltage and Current Capabilities

     

    5.2.1. Test Overview

     

    For this test I am going to use my DC electronic load which is a TTi LD300. This can support up to 80V input and up to 80A within a total power dissipation of 300W. For testing the outputs of the Keysight E36313A I will use the 80A constant current range to enable me to load Channel 1 of the power supply to 10A and I will use the 8A constant current range to load Channel 2 and Channel 3 of the power supply in their various independent, series, and parallel modes.

     

    The following tests are going to be performed on the Keysight E36313A only as we can assume that the TTi units perform to their load specs. I have used these for a number of years and I know they work as advertised in this respect.

     

    5.2.2. Set Channel 1 to 6V and Load with an Increasing Current up to 10A

     

    Keysight E36313A ConfigurationKeysight E36313A Results

     

    The voltage of the channel was set to 6V and the current limit was set to 10A. Initially the output was unloaded and the voltage was measured using my Keysight U1272A Digital Multi Meter. The following table shows the results obtained from varying the load from 0A up to 10A.

     

    Condition
    Result
    6V, No LoadPass - Measured 6.003V on DMM
    6V, 1A LoadPass
    6V, 2A LoadPass
    6V, 5A LoadPass
    6V, 10A LoadPass - Measured 5.895V on DMM
    6V, Over LoadPass - Measured 1V on the DMM

    5.2.3. Set Channel 2 and Channel 3 to 25V and Load Each with an Increasing Current up to 2A

     

    Keysight E36313A ConfigurationKeysight E36313A Results

     

    The voltage of the channel was set to 25V and the current limit was set to 2A. Initially the output was unloaded and the voltage was measured using my Keysight U1272A Digital Multi Meter. The following table shows the results obtained from varying the load from 0A up to 2A.

     

    Condition
    Result
    25V, No LoadPass - Measured 25.013V on DMM
    25V, 200mA LoadPass
    25V, 500mA LoadPass
    25V, 1A LoadPass
    25V, 2A LoadPass - Measured 24.994V on DMM
    25V, Over LoadPass - Measured 0.175V on the DMM

     

    The voltage of the channel was set to 25V and the current limit was set to 2A. Initially the output was unloaded and the voltage was measured using my Keysight U1272A Digital Multi Meter. The following table shows the results obtained from varying the load from 0A up to 2A.

     

    Condition
    Result
    25V, No LoadPass - Measured 25.012V on DMM
    25V, 200mA LoadPass
    25V, 500mA LoadPass
    25V, 1A LoadPass
    25V, 2A LoadPass - Measured 24.994V on DMM
    25V, Over LoadPass - Measured 0.174V on the DMM

     

    5.2.4. Set Channels 2 and 3 into Serial Mode and to 50V and Load with an Increasing Load up to 2A

     

    Keysight E36313A ConfigurationKeysight E36313A Results

     

    Channels 2 and 3 were put into serial mode and the voltage of the combined channel was set to 50V and the current limit was set to 2A. Initially the output was unloaded and the voltage was measured using my Keysight U1272A Digital Multi Meter. The following table shows the results obtained from varying the load from 0A up to 2A.

     

    Header 1Header 2
    50V, No LoadPass - Measured 50.01V on DMM
    50V, 200mA LoadPass
    50V, 500mA LoadPass
    50V, 1A LoadPass
    50V, 2A LoadPass - Measured 49.97 on DMM
    50V, Over LoadPass - Measured 24.7 on DMM

     

    5.2.5. Set Channels 2 and 3 into Parallel Mode and to 25V and Load with an Increasing Load up to 4A

     

    Keysight E36313A ConfigurationKeysight E36313A Results

     

    Channels 2 and 3 were put into parallel mode and the voltage of the combined channel was set to 25 and the current limit was set to 4A. Initially the output was unloaded and the voltage was measured using my Keysight U1272A Digital Multi Meter. The following table shows the results obtained from varying the load from 0A up to 4A.

     

    Header 1Header 2
    25V, No LoadPass - Measured 25.013V on DMM
    25V, 500mA LoadPass
    25V, 1A LoadPass
    25V, 2A LoadPass
    25V, 4A LoadPass - Measured 24.945V on DMM
    25V, Over LoadPass - Measured 0.379V on DMM

     

    5.2.6. Analysis of the Results

     

    Testing under various loads with the maximum voltages available on each channel and the Keysight E36313A didn't miss a beat. It works as expected and was very easy to reconfigure for each test.

     

    5.3. Looking at the regulation under load, light load and transient load conditions

     

    5.3.1. Test Overview

     

    This this test I am just going to compare the Channel 1 output of the Keysight E36313A with the TTi CPX400S and the Channel 2 output with Channel 1 of the TTi PL303QMD. I'll take a look at the ripple and stability of each of the outputs on my Keysight MSOX3054A Oscilloscope using an AC coupled voltage probe when loaded up the outputs with my TTi LD300 electronic load. I will used the web interface available with the LAN Option for the Oscilloscope to capture the resulting waveforms.

     

    5.3.2. Set Channel 1 of the Keysight E36313A and the CPX400S to 5V with a 10A Current Limit and Set the Load to 1A.

     

    Keysight E36313A ConfigurationKeysight E36313A ResultsTTi CPX400S Results  

     

    5.3.3. Set Channel 1 of the Keysight E36313A and the CPX400S to 5V with a 10A Current Limit and Set the Load to 10A.

     

    Keysight E36313A ConfigurationKeysight E36313A ResultsTTi CPX400S Results  

     

    5.3.4. Set Channel 1 of the Keysight E36313A and the CPX400S to 5V with a 10A Current Limit and Set a Step Load from 1A to 10A.

     

    Keysight E36313A ConfigurationKeysight E36313A ResultsTTi CPX400S Results  

     

    5.3.5. Set Channel 2 of the Keysight E36313A and Channel 1 of the PL303QMD to 25V with a 2A Current Limit and Set the Load to 200mA.

     

    Keysight E36313A ConfigurationKeysight E36313A ResultsTTi PL303QMD Results

     

    5.3.6. Set Channel 2 of the Keysight E36313A and Channel 1 of the PL303QMD to 25V with a 2A Current Limit and Set the Load to 2A.

     

    Keysight E36313A ConfigurationKeysight E36313A ResultsTTi PL303QMD Results

     

    5.3.7. Set Channel 2 of the Keysight E36313A and Channel 1 of the PL303QMD to 15V with a 2A Current Limit and Set a Step Load from 200mA to 2A.

     

    Keysight E36313A ConfigurationKeysight E36313A ResultsTTi PL303QMD Results

     

    5.3.8. Analysis of the Results

     

    We can see from the results that all three supplies are very good. The Keysight E36313A has strong performance in all the tests, beating the TTi CPX400S in the transient load and output noise measurements, where the TTi CPX400S had some noticeable regular switching spikes which increased with current. At the higher voltage and lower current tests the Keysight E36313A performed roughly on a par with the TTi PL303QMD, having slightly more excursion low but slightly less excursion high on the transient voltage test.

     

    Under 200ma/2A transient load test the TTi power supply would run but would occasionally trip even though I had actually set the current limit to a little over 2A. They Keysight E36313A did not trip out during this test at all.

     

    5.4. FFT analysis of the fully loaded output

     

    5.4.1. Test Overview

     

    For this test I will look at just the single output with 1A and 10A load conditions setup from the previous tests and using the FFT function of my Keysight MSOX3054A I will attempt to see if there are any noticeable peaks in the spectrum which are dependant on load which are detectable within the capabilities of my oscilloscope. I will used the web interface available with the LAN Option for the Oscilloscope to capture the resulting waveforms.

     

    5.4.2. Set Channel 1 of the Keysight E36313A and the CPX400S to 5V with a 10A Current Limit and Set the Load to 1A.

     

    Keysight E36313A ConfigurationKeysight E36313A ResultsTTi CPX400S Results  

     

    5.4.3. Set Channel 1 of the Keysight E36313A and the CPX400S to 5V with a 10A Current Limit and Set the Load to 10A.

     

    Keysight E36313A ConfigurationKeysight E36313A ResultsTTi CPX400S Results  

     

    5.4.4. Analysis of the Results

     

    Looking at the results we can see that the TTi CPX400S has got a higher peak at 125MHz and additional spurs at 187.5MHz, 200MHz, 212.5MHz and 225MHz which are either not present or at a much lower level on the output of the Keysight E36313A. These likely correlate with the edge speed of the switching noise we saw in the previous output noise tests but without further investigation I cannot be sure.

     

    We can also see from the FFT that there are spurs in the 85MHz to 105MHz range for all tests and all equipment. These are present even under no load and with the output disabled on both units, I think these are actually artefacts which are caused by some digital data being picked up from elsewhere. I haven't tracked down from exactly where yet but if we look at the following images you can see the regularity of the bursts and with the zoomed image we can see something which looks like it might be in the 20MHz-25MHz range. It's obviously very hard to see the details of the data streams from a spurious waveform captured on my oscilloscope.

     

     

    5.5. Testing out the data logging

    5.5.1. Test Overview

     

    For this test I will be capturing a set of data over time and recording it to a file which I will then import into a spreadsheet application to perform some graphing of the results. I have chosen to log the Voltage, Current and Power measurements when booting up a Raspberry Pi Model B and running some benchmark tests. For my tests I will be performing 5 separate actions

     

    1. Booting up the Raspberry Pi from powered off and logging in.

    2. Running a sysbench benchmark on the CPU to calculate prime numbers.

      sysbench --test=cpu --cpu-max-prime=2000 run
    3. Running a sysbench benchmark on the Memory to write 2GB of data.

      sysbench --test=memory run --memory-total-size=2G
    4. Running a sysbench benchmark on the Memory to read 2GB of data.

      sysbench --test=memory run --memory-total-size=2G --memory-oper=read
    5. Shutting down and powering off the Raspberry Pi.

      sudo shutdown -HP now

     

    5.5.2. Performing the Tests

     

    The unit comes with a relatively straight forward to use data logger. You can configure what values are captured, i.e. V/I/P from Channels 1/2/3, you can configure how long it captures for, how often it samples the data etc. As it's logging it will show you the waveform of the log in a graph which is like a little oscilloscope with a slow time base. You need a USB stick to capture the data and you can save it in a binary dlog format and also in CSV format.

     

    5.5.3. Test Data Collected

     

    The following image shows the log view of the data collected:

     

    I also captured the log data into a CSV file:

     

    E36313A exported datalog
    File type: CSV
    
    
    "Sample interval: 0.200000"
    "Number of sample: 1026/1650"
    "Trigger sample: 0"
    "Date: 2017/12/19 14:28:28"
    
    
    "V1: 5.000000, -7.500000"
    "I1: 0.200000, 0.150000"
    "P1: 1.000000, 2.781250"
    "V2: 5.000000, 0.000000"
    "I2: 1.000000, 0.000000"
    "P2: 2.000000, 0.000000"
    "V3: 5.000000, 0.000000"
    "I3: 1.000000, 0.000000"
    "P3: 2.000000, 0.000000"
    
    
    Sample,"Volt avg 1","Curr avg 1"
    0,0.001310,0.000070
    1,0.001310,0.000070
    2,0.001310,0.000070
    3,0.001310,0.000070
    4,0.001310,0.000070
    5,0.001310,0.000070
    6,0.001310,0.000070
    7,0.001310,0.000070
    8,4.999190,0.084240
    9,5.000390,0.113180
    10,5.000590,0.095300
    11,5.000590,0.076760
    12,5.000590,0.102770
    13,5.000590,0.098550
    14,5.000590,0.067660
    15,5.000590,0.098220
    16,5.000590,0.104070
    17,5.000590,0.107650
    18,5.000590,0.082940
    19,5.000790,0.067330
    20,5.000590,0.067330
    21,5.000590,0.067330
    22,5.000590,0.067330
    23,5.000590,0.074490
    24,5.000590,0.127810
    25,5.000590,0.127810
    26,5.000590,0.128130
    27,5.000590,0.121630
    28,5.000590,0.141140
    29,5.000590,0.111550
    30,5.000590,0.149920
    31,5.000590,0.156750
    32,5.000590,0.157070
    33,5.000590,0.115780
    34,5.000590,0.115780
    35,5.000590,0.115780
    36,5.000590,0.115780
    37,5.000590,0.187960
    38,5.000590,0.227950
    39,5.000790,0.194140
    40,5.000590,0.242910
    41,5.000590,0.259490
    42,5.000590,0.281270
    43,5.000590,0.288100
    44,5.000590,0.277040
    45,5.000590,0.278350
    46,5.000590,0.277690
    47,5.000590,0.278020
    48,5.000590,0.262410
    49,5.000590,0.323210
    50,5.000590,0.279970
    51,5.000590,0.280950
    52,5.000790,0.307280
    53,5.000790,0.280950
    54,5.000590,0.269570
    55,5.000790,0.271190
    56,5.000590,0.270220
    57,5.000790,0.255590
    58,5.000590,0.260460
    59,5.000590,0.267940
    60,5.000790,0.280620
    61,5.000790,0.250060
    62,5.000790,0.264690
    63,5.000590,0.261110
    64,5.000590,0.251360
    65,5.000590,0.251360
    66,5.000590,0.251030
    67,5.000790,0.258190
    68,5.000790,0.249730
    69,5.000590,0.262740
    70,5.000590,0.261440
    71,5.000790,0.253640
    72,5.000590,0.259810
    73,5.000590,0.278350
    74,5.000790,0.240300
    75,5.000590,0.251030
    76,5.000790,0.266640
    77,5.000590,0.253960
    78,5.000590,0.258190
    79,5.000590,0.254940
    80,5.000790,0.267290
    81,5.000590,0.251030
    82,5.000590,0.259160
    83,5.000590,0.246810
    84,5.000590,0.262090
    85,5.000590,0.252330
    86,5.000790,0.257210
    87,5.000790,0.252010
    88,5.000790,0.251360
    89,5.000790,0.253640
    90,5.000590,0.254940
    91,5.000590,0.245830
    92,5.000590,0.235430
    93,5.000590,0.234780
    94,5.000790,0.235430
    95,5.000590,0.237380
    96,5.000590,0.271520
    97,5.000590,0.247130
    98,5.000590,0.235430
    99,5.000590,0.235430
    100,5.000790,0.252010
    101,5.000590,0.236730
    102,5.000790,0.250060
    103,5.000790,0.252010
    104,5.000790,0.260790
    105,5.000790,0.250060
    106,5.000590,0.254610
    107,5.000590,0.256890
    108,5.000590,0.278020
    109,5.000790,0.257210
    110,5.000590,0.245830
    111,5.000790,0.258840
    112,5.000590,0.250380
    113,5.000590,0.252010
    114,5.000590,0.238030
    115,5.000590,0.249730
    116,5.000590,0.254940
    117,5.000790,0.239980
    118,5.000590,0.252980
    119,5.000790,0.250380
    120,5.000590,0.236400
    121,5.000590,0.251030
    122,5.000790,0.268920
    123,5.000590,0.250380
    124,5.000590,0.259490
    125,5.000790,0.261440
    126,5.000590,0.252660
    127,5.000590,0.242580
    128,5.000590,0.252980
    129,5.000590,0.254940
    130,5.000590,0.258190
    131,5.000590,0.260790
    132,5.000590,0.257210
    133,5.000790,0.240950
    134,5.000790,0.235100
    135,5.000590,0.250380
    136,5.000590,0.240630
    137,5.000590,0.251360
    138,5.000590,0.253960
    139,5.000590,0.260460
    140,5.000790,0.251680
    141,5.000790,0.250710
    142,5.000590,0.241280
    143,5.000590,0.272170
    144,5.000590,0.235430
    145,5.000790,0.235100
    146,5.000590,0.235430
    147,5.000590,0.237050
    148,5.000590,0.240300
    149,5.000590,0.235100
    150,5.000590,0.248110
    151,5.000590,0.248760
    152,5.000590,0.235100
    153,5.000790,0.248430
    154,5.000590,0.268590
    155,5.000590,0.269570
    156,5.000790,0.227950
    157,5.000590,0.211040
    158,5.000790,0.210720
    159,5.000590,0.210720
    160,5.000590,0.210720
    161,5.000590,0.210720
    162,5.000590,0.210720
    163,5.000590,0.210720
    164,5.000590,0.210720
    165,5.000590,0.211040
    166,5.000590,0.210720
    167,5.000590,0.211040
    168,5.000590,0.210720
    169,5.000590,0.210720
    170,5.000590,0.210390
    171,5.000590,0.210720
    172,5.000590,0.210390
    173,5.000590,0.210720
    174,5.000590,0.211040
    175,5.000590,0.210720
    176,5.000590,0.210720
    177,5.000590,0.211040
    178,5.000590,0.210720
    179,5.000790,0.210720
    180,5.000790,0.210390
    181,5.000590,0.210390
    182,5.000590,0.210720
    183,5.000590,0.210720
    184,5.000590,0.210720
    185,5.000590,0.211040
    186,5.000590,0.211370
    187,5.000590,0.210720
    188,5.000590,0.211370
    189,5.000590,0.210720
    190,5.000590,0.210390
    191,5.000590,0.280620
    192,5.000590,0.279970
    193,5.000590,0.279650
    194,5.000590,0.281270
    195,5.000590,0.280620
    196,5.000590,0.279320
    197,5.000590,0.280620
    198,5.000590,0.279000
    199,5.000590,0.279000
    200,5.000590,0.225020
    201,5.000590,0.210390
    202,5.000590,0.211040
    203,5.000790,0.210720
    204,5.000590,0.210720
    205,5.000590,0.210720
    206,5.000590,0.211040
    207,5.000590,0.210720
    208,5.000590,0.210390
    209,5.000590,0.210720
    210,5.000590,0.210720
    211,5.000590,0.210720
    212,5.000590,0.211040
    213,5.000590,0.210720
    214,5.000590,0.210720
    215,5.000590,0.210720
    216,5.000590,0.210720
    217,5.000590,0.210720
    218,5.000590,0.210720
    219,5.000590,0.210720
    220,5.000590,0.210720
    221,5.000590,0.210720
    222,5.000590,0.211040
    223,5.000590,0.211040
    224,5.000590,0.210720
    225,5.000590,0.211040
    226,5.000590,0.210720
    227,5.000590,0.210390
    228,5.000590,0.210720
    229,5.000790,0.210720
    230,5.000790,0.210720
    231,5.000790,0.210720
    232,5.000790,0.210720
    233,5.000790,0.210720
    234,5.000790,0.210720
    235,5.000790,0.210720
    236,5.000790,0.210720
    237,5.000590,0.210720
    238,5.000590,0.210720
    239,5.000590,0.210720
    240,5.000590,0.210720
    241,5.000590,0.281920
    242,5.000590,0.279970
    243,5.000590,0.279650
    244,5.000590,0.279320
    245,5.000590,0.279320
    246,5.000590,0.279320
    247,5.000590,0.279320
    248,5.000590,0.279320
    249,5.000590,0.279320
    250,5.000590,0.279320
    251,5.000590,0.279320
    252,5.000590,0.211040
    253,5.000590,0.211040
    254,5.000590,0.210720
    255,5.000590,0.211040
    256,5.000590,0.211040
    257,5.000590,0.211040
    258,5.000590,0.210720
    259,5.000590,0.210720
    260,5.000590,0.210720
    261,5.000590,0.210720
    262,5.000590,0.210720
    263,5.000590,0.210720
    264,5.000590,0.210720
    265,5.000590,0.210720
    266,5.000590,0.214940
    267,5.000790,0.224700
    268,5.000590,0.210720
    269,5.000590,0.210720
    270,5.000590,0.210720
    271,5.000590,0.210720
    272,5.000590,0.211040
    273,5.000590,0.210720
    274,5.000590,0.210720
    275,5.000590,0.210720
    276,5.000590,0.210720
    277,5.000590,0.210720
    278,5.000590,0.210720
    279,5.000590,0.210720
    280,5.000590,0.210720
    281,5.000590,0.210720
    282,5.000590,0.211370
    283,5.000590,0.253960
    284,5.000590,0.241930
    285,5.000590,0.305980
    286,5.000590,0.305980
    287,5.000590,0.262740
    288,5.000590,0.212340
    289,5.000590,0.212340
    290,5.000590,0.212340
    291,5.000590,0.212340
    292,5.000590,0.212340
    293,5.000590,0.212340
    294,5.000590,0.212340
    295,5.000590,0.212340
    296,5.000590,0.212340
    297,5.000590,0.279650
    298,5.000590,0.281270
    299,5.000590,0.279970
    300,5.000590,0.279320
    301,5.000590,0.225020
    302,5.000590,0.211040
    303,5.000590,0.211370
    304,5.000590,0.211040
    305,5.000590,0.211040
    306,5.000590,0.211040
    307,5.000590,0.211040
    308,5.000590,0.211040
    309,5.000590,0.211040
    310,5.000590,0.211040
    311,5.000590,0.211040
    312,5.000590,0.225350
    313,5.000590,0.211040
    314,5.000590,0.210720
    315,5.000790,0.211040
    316,5.000590,0.210720
    317,5.000590,0.211370
    318,5.000590,0.211040
    319,5.000590,0.211040
    320,5.000590,0.211040
    321,5.000590,0.211040
    322,5.000590,0.211040
    323,5.000590,0.211040
    324,5.000590,0.211040
    325,5.000590,0.211040
    326,5.000590,0.225670
    327,5.000590,0.210720
    328,5.000590,0.210720
    329,5.000590,0.211040
    330,5.000590,0.210720
    331,5.000590,0.211370
    332,5.000590,0.211370
    333,5.000590,0.210720
    334,5.000590,0.210390
    335,5.000590,0.210390
    336,5.000590,0.210390
    337,5.000590,0.210390
    338,5.000590,0.210390
    339,5.000590,0.210390
    340,5.000590,0.210390
    341,5.000590,0.210390
    342,5.000590,0.248760
    343,5.000790,0.261110
    344,5.000790,0.269240
    345,5.000590,0.250060
    346,5.000590,0.235430
    347,5.000590,0.250710
    348,5.000790,0.251030
    349,5.000590,0.251030
    350,5.000590,0.251030
    351,5.000590,0.251030
    352,5.000590,0.251030
    353,5.000590,0.251030
    354,5.000590,0.251030
    355,5.000590,0.251030
    356,5.000590,0.251030
    357,5.000590,0.210720
    358,5.000590,0.211040
    359,5.000790,0.210720
    360,5.000590,0.210720
    361,5.000590,0.211040
    362,5.000590,0.210720
    363,5.000590,0.211370
    364,5.000590,0.211370
    365,5.000590,0.211370
    366,5.000590,0.211370
    367,5.000590,0.211370
    368,5.000590,0.211370
    369,5.000590,0.211370
    370,5.000590,0.211370
    371,5.000590,0.210720
    372,5.000590,0.210720
    373,5.000590,0.210720
    374,5.000590,0.210720
    375,5.000590,0.225020
    376,5.000590,0.211370
    377,5.000590,0.211040
    378,5.000590,0.211040
    379,5.000590,0.211040
    380,5.000590,0.211040
    381,5.000590,0.211040
    382,5.000590,0.211040
    383,5.000590,0.211040
    384,5.000590,0.211040
    385,5.000590,0.211040
    386,5.000790,0.241610
    387,5.000590,0.241280
    388,5.000590,0.241610
    389,5.000590,0.241280
    390,5.000590,0.241280
    391,5.000590,0.241280
    392,5.000790,0.241610
    393,5.000590,0.241610
    394,5.000590,0.241610
    395,5.000590,0.241610
    396,5.000590,0.241610
    397,5.000590,0.241610
    398,5.000590,0.241610
    399,5.000590,0.241610
    400,5.000590,0.241610
    401,5.000590,0.241610
    402,5.000590,0.241280
    403,5.000590,0.241280
    404,5.000590,0.241280
    405,5.000590,0.241280
    406,5.000590,0.241280
    407,5.000590,0.241280
    408,5.000590,0.241280
    409,5.000590,0.241280
    410,5.000590,0.241280
    411,5.000590,0.241280
    412,5.000590,0.241610
    413,5.000590,0.241610
    414,5.000790,0.241610
    415,5.000590,0.241610
    416,5.000590,0.241280
    417,5.000790,0.241280
    418,5.000590,0.311830
    419,5.000590,0.311830
    420,5.000590,0.311830
    421,5.000590,0.311830
    422,5.000590,0.311830
    423,5.000590,0.311830
    424,5.000590,0.311830
    425,5.000590,0.311830
    426,5.000590,0.250060
    427,5.000590,0.241280
    428,5.000590,0.241610
    429,5.000590,0.241610
    430,5.000590,0.241610
    431,5.000590,0.241610
    432,5.000590,0.241610
    433,5.000590,0.241280
    434,5.000590,0.241280
    435,5.000590,0.241280
    436,5.000590,0.241280
    437,5.000590,0.241280
    438,5.000590,0.241280
    439,5.000590,0.241280
    440,5.000590,0.241280
    441,5.000590,0.241610
    442,5.000590,0.241610
    443,5.000590,0.245830
    444,5.000590,0.241610
    445,5.000590,0.241610
    446,5.000590,0.241610
    447,5.000590,0.241610
    448,5.000590,0.241280
    449,5.000590,0.241280
    450,5.000590,0.241280
    451,5.000590,0.241280
    452,5.000590,0.241280
    453,5.000590,0.241280
    454,5.000590,0.241280
    455,5.000590,0.241280
    456,5.000590,0.241280
    457,5.000590,0.241280
    458,5.000590,0.241610
    459,5.000590,0.241610
    460,5.000590,0.241610
    461,5.000590,0.241610
    462,5.000590,0.241610
    463,5.000590,0.241280
    464,5.000590,0.241280
    465,5.000590,0.241280
    466,5.000590,0.241280
    467,5.000590,0.241280
    468,5.000590,0.241280
    469,5.000590,0.241280
    470,5.000590,0.241280
    471,5.000590,0.241280
    472,5.000590,0.241610
    473,5.000590,0.241610
    474,5.000590,0.241610
    475,5.000790,0.241610
    476,5.000590,0.241280
    477,5.000590,0.241610
    478,5.000590,0.241610
    479,5.000590,0.241610
    480,5.000590,0.241610
    481,5.000590,0.241610
    482,5.000590,0.241610
    483,5.000590,0.241610
    484,5.000590,0.241610
    485,5.000590,0.241610
    486,5.000590,0.241280
    487,5.000590,0.241280
    488,5.000590,0.241610
    489,5.000590,0.241610
    490,5.000590,0.241610
    491,5.000590,0.241610
    492,5.000590,0.241610
    493,5.000590,0.312160
    494,5.000590,0.312160
    495,5.000590,0.312160
    496,5.000590,0.312160
    497,5.000590,0.312160
    498,5.000590,0.312160
    499,5.000590,0.312160
    500,5.000590,0.312160
    501,5.000590,0.241610
    502,5.000590,0.241610
    503,5.000590,0.241610
    504,5.000790,0.241610
    505,5.000590,0.241610
    506,5.000590,0.241610
    507,5.000590,0.241610
    508,5.000590,0.241280
    509,5.000590,0.241280
    510,5.000590,0.241280
    511,5.000590,0.241280
    512,5.000590,0.241280
    513,5.000590,0.241280
    514,5.000590,0.241280
    515,5.000590,0.241280
    516,5.000590,0.241610
    517,5.000590,0.238030
    518,5.000590,0.238350
    519,5.000590,0.241610
    520,5.000590,0.241610
    521,5.000590,0.241610
    522,5.000590,0.241280
    523,5.000590,0.241610
    524,5.000590,0.241610
    525,5.000590,0.241610
    526,5.000590,0.241610
    527,5.000590,0.241610
    528,5.000590,0.241610
    529,5.000590,0.241610
    530,5.000590,0.241610
    531,5.000590,0.241610
    532,5.000590,0.241610
    533,5.000590,0.241280
    534,5.000590,0.241610
    535,5.000590,0.241280
    536,5.000590,0.241610
    537,5.000590,0.241610
    538,5.000590,0.241610
    539,5.000590,0.241610
    540,5.000590,0.241610
    541,5.000590,0.241610
    542,5.000590,0.241610
    543,5.000590,0.241610
    544,5.000590,0.241610
    545,5.000590,0.241610
    546,5.000590,0.241610
    547,5.000590,0.241610
    548,5.000590,0.255590
    549,5.000590,0.241610
    550,5.000790,0.241610
    551,5.000590,0.241610
    552,5.000590,0.241280
    553,5.000590,0.241280
    554,5.000590,0.241280
    555,5.000590,0.241280
    556,5.000590,0.241280
    557,5.000590,0.241280
    558,5.000590,0.241280
    559,5.000590,0.241280
    560,5.000590,0.241280
    561,5.000590,0.241610
    562,5.000590,0.241280
    563,5.000590,0.264360
    564,5.000590,0.263710
    565,5.000590,0.225020
    566,5.000590,0.210720
    567,5.000590,0.210720
    568,5.000590,0.210720
    569,5.000590,0.210720
    570,5.000590,0.210720
    571,5.000590,0.210720
    572,5.000590,0.210720
    573,5.000590,0.210720
    574,5.000590,0.210720
    575,5.000590,0.210720
    576,5.000590,0.210720
    577,5.000590,0.232180
    578,5.000590,0.210720
    579,5.000590,0.212020
    580,5.000590,0.210720
    581,5.000590,0.211040
    582,5.000590,0.211040
    583,5.000590,0.210720
    584,5.000590,0.210720
    585,5.000590,0.210720
    586,5.000590,0.210720
    587,5.000590,0.210720
    588,5.000590,0.210720
    589,5.000590,0.210720
    590,5.000590,0.210720
    591,5.000590,0.210720
    592,5.000790,0.211040
    593,5.000590,0.210720
    594,5.000590,0.210720
    595,5.000590,0.210720
    596,5.000590,0.210720
    597,5.000590,0.210720
    598,5.000590,0.210720
    599,5.000590,0.210720
    600,5.000590,0.210720
    601,5.000590,0.210720
    602,5.000590,0.210720
    603,5.000590,0.210720
    604,5.000590,0.210720
    605,5.000590,0.210720
    606,5.000590,0.247130
    607,5.000590,0.246810
    608,5.000590,0.247130
    609,5.000590,0.247130
    610,5.000590,0.246810
    611,5.000590,0.247130
    612,5.000590,0.247130
    613,5.000590,0.246810
    614,5.000590,0.246810
    615,5.000590,0.246810
    616,5.000590,0.246810
    617,5.000590,0.246810
    618,5.000590,0.246810
    619,5.000590,0.246810
    620,5.000590,0.246810
    621,5.000590,0.246810
    622,5.000590,0.247130
    623,5.000590,0.247130
    624,5.000590,0.247130
    625,5.000590,0.246810
    626,5.000590,0.247130
    627,5.000590,0.247130
    628,5.000590,0.247130
    629,5.000590,0.247130
    630,5.000590,0.247130
    631,5.000590,0.247130
    632,5.000590,0.247130
    633,5.000590,0.247130
    634,5.000590,0.247130
    635,5.000590,0.247130
    636,5.000590,0.246810
    637,5.000590,0.247130
    638,5.000590,0.247130
    639,5.000590,0.247130
    640,5.000790,0.247130
    641,5.000590,0.247130
    642,5.000590,0.247130
    643,5.000590,0.246810
    644,5.000590,0.247130
    645,5.000590,0.247130
    646,5.000590,0.247130
    647,5.000590,0.247130
    648,5.000590,0.247130
    649,5.000590,0.247130
    650,5.000590,0.247130
    651,5.000590,0.247130
    652,5.000590,0.247130
    653,5.000590,0.247130
    654,5.000790,0.247130
    655,5.000590,0.247130
    656,5.000790,0.247130
    657,5.000590,0.247130
    658,5.000590,0.247130
    659,5.000590,0.247130
    660,5.000590,0.247130
    661,5.000590,0.247130
    662,5.000590,0.247130
    663,5.000590,0.247130
    664,5.000590,0.247130
    665,5.000590,0.247130
    666,5.000590,0.247130
    667,5.000590,0.247130
    668,5.000790,0.247130
    669,5.000590,0.247460
    670,5.000590,0.317360
    671,5.000590,0.316060
    672,5.000590,0.315740
    673,5.000590,0.261110
    674,5.000590,0.247130
    675,5.000590,0.247130
    676,5.000590,0.247130
    677,5.000590,0.247130
    678,5.000590,0.247130
    679,5.000590,0.247130
    680,5.000590,0.247130
    681,5.000590,0.247130
    682,5.000590,0.246810
    683,5.000590,0.246810
    684,5.000590,0.247130
    685,5.000590,0.247130
    686,5.000590,0.247130
    687,5.000590,0.247130
    688,5.000590,0.247130
    689,5.000590,0.247130
    690,5.000590,0.247130
    691,5.000590,0.247130
    692,5.000590,0.247130
    693,5.000590,0.247130
    694,5.000590,0.247130
    695,5.000590,0.247130
    696,5.000590,0.247130
    697,5.000590,0.247130
    698,5.000590,0.249410
    699,5.000590,0.247130
    700,5.000590,0.247130
    701,5.000590,0.247130
    702,5.000590,0.247130
    703,5.000590,0.247130
    704,5.000590,0.247130
    705,5.000590,0.247130
    706,5.000590,0.247130
    707,5.000590,0.247130
    708,5.000590,0.247130
    709,5.000590,0.247130
    710,5.000590,0.247130
    711,5.000590,0.247130
    712,5.000590,0.247130
    713,5.000590,0.247130
    714,5.000590,0.247130
    715,5.000590,0.247130
    716,5.000590,0.247130
    717,5.000590,0.247130
    718,5.000590,0.247130
    719,5.000590,0.247130
    720,5.000590,0.247130
    721,5.000590,0.247130
    722,5.000590,0.247130
    723,5.000590,0.247130
    724,5.000590,0.247130
    725,5.000590,0.247130
    726,5.000590,0.247130
    727,5.000590,0.247130
    728,5.000590,0.262740
    729,5.000590,0.242910
    730,5.000590,0.211040
    731,5.000590,0.210720
    732,5.000790,0.210720
    733,5.000590,0.210720
    734,5.000590,0.210720
    735,5.000590,0.210720
    736,5.000590,0.210720
    737,5.000590,0.210720
    738,5.000590,0.210720
    739,5.000590,0.210720
    740,5.000590,0.210720
    741,5.000590,0.210720
    742,5.000590,0.210720
    743,5.000590,0.210720
    744,5.000590,0.210720
    745,5.000590,0.211040
    746,5.000590,0.211040
    747,5.000590,0.210720
    748,5.000590,0.210720
    749,5.000590,0.210720
    750,5.000590,0.210720
    751,5.000590,0.210720
    752,5.000590,0.210720
    753,5.000590,0.210720
    754,5.000590,0.210720
    755,5.000590,0.210720
    756,5.000590,0.210720
    757,5.000590,0.211040
    758,5.000590,0.210720
    759,5.000590,0.210720
    760,5.000590,0.210720
    761,5.000590,0.210390
    762,5.000590,0.210720
    763,5.000590,0.211040
    764,5.000590,0.211040
    765,5.000590,0.211040
    766,5.000590,0.211040
    767,5.000590,0.211040
    768,5.000590,0.211040
    769,5.000590,0.211040
    770,5.000590,0.211040
    771,5.000590,0.211040
    772,5.000590,0.210720
    773,5.000590,0.211040
    774,5.000590,0.211040
    775,5.000590,0.210720
    776,5.000590,0.210720
    777,5.000590,0.276070
    778,5.000590,0.234780
    779,5.000590,0.234780
    780,5.000590,0.234780
    781,5.000590,0.234780
    782,5.000590,0.234780
    783,5.000590,0.234780
    784,5.000590,0.234780
    785,5.000590,0.234780
    786,5.000590,0.234780
    787,5.000590,0.234780
    788,5.000590,0.234780
    789,5.000590,0.234780
    790,5.000590,0.234780
    791,5.000590,0.234780
    792,5.000590,0.235100
    793,5.000590,0.234780
    794,5.000590,0.234780
    795,5.000590,0.234780
    796,5.000590,0.234780
    797,5.000590,0.234780
    798,5.000590,0.234780
    799,5.000590,0.234780
    800,5.000590,0.234780
    801,5.000590,0.235100
    802,5.000590,0.234780
    803,5.000590,0.234780
    804,5.000590,0.234780
    805,5.000590,0.234780
    806,5.000590,0.234780
    807,5.000590,0.234780
    808,5.000590,0.234780
    809,5.000590,0.234780
    810,5.000590,0.234780
    811,5.000590,0.234780
    812,5.000590,0.234780
    813,5.000590,0.234780
    814,5.000590,0.234780
    815,5.000590,0.234780
    816,5.000590,0.234780
    817,5.000590,0.234780
    818,5.000590,0.234780
    819,5.000590,0.234780
    820,5.000590,0.234780
    821,5.000590,0.234780
    822,5.000590,0.234780
    823,5.000590,0.234780
    824,5.000590,0.234780
    825,5.000590,0.234780
    826,5.000590,0.234780
    827,5.000590,0.234780
    828,5.000590,0.234780
    829,5.000590,0.235100
    830,5.000590,0.234780
    831,5.000590,0.234780
    832,5.000590,0.234780
    833,5.000590,0.234780
    834,5.000590,0.234780
    835,5.000590,0.234780
    836,5.000590,0.234780
    837,5.000590,0.234780
    838,5.000590,0.234780
    839,5.000590,0.234780
    840,5.000590,0.234780
    841,5.000590,0.234780
    842,5.000590,0.234780
    843,5.000590,0.234780
    844,5.000590,0.234780
    845,5.000590,0.234780
    846,5.000590,0.234780
    847,5.000590,0.234780
    848,5.000590,0.234780
    849,5.000590,0.234780
    850,5.000590,0.234780
    851,5.000590,0.234780
    852,5.000590,0.234780
    853,5.000590,0.234780
    854,5.000590,0.234780
    855,5.000590,0.234780
    856,5.000590,0.234780
    857,5.000790,0.234780
    858,5.000590,0.234780
    859,5.000590,0.234780
    860,5.000590,0.234780
    861,5.000590,0.234780
    862,5.000590,0.234780
    863,5.000590,0.234780
    864,5.000590,0.234780
    865,5.000590,0.234780
    866,5.000590,0.234780
    867,5.000590,0.234780
    868,5.000590,0.234780
    869,5.000590,0.234780
    870,5.000590,0.234780
    871,5.000590,0.234780
    872,5.000590,0.234780
    873,5.000590,0.234780
    874,5.000590,0.234780
    875,5.000590,0.234780
    876,5.000590,0.234780
    877,5.000590,0.234780
    878,5.000590,0.235100
    879,5.000590,0.234780
    880,5.000590,0.234780
    881,5.000590,0.234780
    882,5.000590,0.234780
    883,5.000590,0.234780
    884,5.000590,0.234780
    885,5.000590,0.234780
    886,5.000590,0.234780
    887,5.000790,0.210720
    888,5.000590,0.210720
    889,5.000790,0.210720
    890,5.000590,0.211040
    891,5.000590,0.211040
    892,5.000590,0.210720
    893,5.000790,0.210720
    894,5.000790,0.210720
    895,5.000790,0.210720
    896,5.000790,0.210720
    897,5.000790,0.210720
    898,5.000790,0.210720
    899,5.000790,0.210720
    900,5.000790,0.210720
    901,5.000590,0.211040
    902,5.000590,0.211370
    903,5.000790,0.211040
    904,5.000590,0.211040
    905,5.000590,0.210720
    906,5.000590,0.211040
    907,5.000590,0.211690
    908,5.000790,0.210720
    909,5.000790,0.210720
    910,5.000790,0.210720
    911,5.000790,0.210720
    912,5.000790,0.210720
    913,5.000790,0.210720
    914,5.000790,0.210720
    915,5.000790,0.210720
    916,5.000590,0.210720
    917,5.000590,0.210720
    918,5.000590,0.211690
    919,5.000790,0.210720
    920,5.000590,0.210720
    921,5.000590,0.210720
    922,5.000790,0.211040
    923,5.000590,0.210720
    924,5.000790,0.210720
    925,5.000790,0.210720
    926,5.000790,0.210720
    927,5.000790,0.210720
    928,5.000790,0.210720
    929,5.000790,0.210720
    930,5.000790,0.210720
    931,5.000790,0.210720
    932,5.000590,0.211040
    933,5.000590,0.210720
    934,5.000590,0.210720
    935,5.000590,0.250380
    936,5.000590,0.249730
    937,5.000590,0.251360
    938,5.000590,0.236400
    939,5.000590,0.250710
    940,5.000590,0.250710
    941,5.000590,0.250710
    942,5.000590,0.250710
    943,5.000590,0.250710
    944,5.000590,0.250710
    945,5.000590,0.250710
    946,5.000590,0.250710
    947,5.000590,0.235430
    948,5.000590,0.235100
    949,5.000590,0.236730
    950,5.000590,0.275740
    951,5.000590,0.250060
    952,5.000590,0.236730
    953,5.000590,0.250710
    954,5.000590,0.250710
    955,5.000590,0.250710
    956,5.000590,0.250710
    957,5.000590,0.250710
    958,5.000590,0.250710
    959,5.000590,0.250710
    960,5.000590,0.250710
    961,5.000590,0.288750
    962,5.000590,0.244210
    963,5.000590,0.252660
    964,5.000590,0.281600
    965,5.000590,0.279970
    966,5.000590,0.279650
    967,5.000590,0.281270
    968,5.000590,0.279650
    969,5.000590,0.279650
    970,5.000590,0.279650
    971,5.000590,0.279650
    972,5.000590,0.279650
    973,5.000590,0.279650
    974,5.000590,0.279650
    975,5.000590,0.279650
    976,5.000590,0.076440
    977,5.000590,0.073840
    978,5.000590,0.075790
    979,5.000590,0.073510
    980,5.000590,0.075790
    981,5.000590,0.073510
    982,5.000590,0.075460
    983,5.000590,0.073510
    984,5.000590,0.075790
    985,5.000590,0.075790
    986,5.000590,0.075790
    987,5.000590,0.075790
    988,5.000590,0.075790
    989,5.000590,0.075790
    990,5.000590,0.075790
    991,5.000590,0.075790
    992,5.000590,0.075140
    993,5.000590,0.073510
    994,5.000590,0.075460
    995,5.000590,0.073510
    996,5.000590,0.073510
    997,5.000590,0.073510
    998,5.000590,0.073510
    999,5.000590,0.073510
    1000,5.000590,0.073510
    1001,5.000590,0.073510
    1002,5.000590,0.073510
    1003,5.000590,0.073510
    1004,5.000590,0.073510
    1005,5.000590,0.073510
    1006,5.000590,0.071560
    1007,5.000590,0.071560
    1008,5.000590,0.071560
    1009,5.000590,0.071560
    1010,5.000590,0.071560
    1011,5.000590,0.071560
    1012,5.000590,0.071560
    1013,5.000590,0.071560
    1014,5.000590,0.071560
    1015,5.000590,0.071560
    1016,5.000590,0.071560
    1017,5.000590,0.071560
    1018,5.000590,0.071560
    1019,5.000590,0.071560
    1020,5.000590,0.071560
    1021,5.000590,0.071560
    1022,5.000590,0.071560
    1023,5.000590,0.071560
    1024,5.000590,0.071560
    1025,5.000590,0.071560

     

    For some reason it has not included the logged power data in the CSV file. I need to investigate why this is the case. It's not actually an issue as we can reconstruct the power from the V and I data, which I will do in excel with a simple formula.

     

    5.5.4. Using the Results Data

     

    Once you have the data in a useful format there are many things you could so with it depending upon what the data represents. The simplest thing you can do is simply graph the data in Excel or Open/LibreOffice Calc (other Spreadsheets are available too, just use whatever you like and/or are familiar with). The following is a graph of the current (blue trace) and calculated power (red trace) for the experiment of booting up the Pi and running various benchmarks.

     

    Using such tools you could easily calculate the average power consumption of a piece of equipment over time, or correlate particular operations happening on the hardware or in the software with peaks in power consumption, which,  if you are designing something that is required to run on batteries and needs to be low power, this can give some real insights into where to target your efforts in the design to get the best performance for the lowest power consumption.

     

    You could of course bring the data into more comprehensive tools like MatLab, Octave, etc or write your own custom tools to process the data specific for your needs.

     

    UPDATE 21/12/2017:

     

    I've been experimenting and using the logging a little more and it really is very useful and as dougw shows in one of the videos in his road test, there are plenty of more creative ways to make use of the logger. I have found a few limitations though which I think could be improved upon.

     

    Firstly, when you set the logging period for an extended time, it gets to a point where is says there is not enough device memory and it starts to increase the logging interval. Why is this? If I am logging to a say 16Gb USB stick why am I restricted to the log size based upon memory available within the power supply itself? I would have thought even if it needs to use internal memory to capture the initial data, flushing this data to a file periodically should allow this to be able to continuously log until the USB is full. I hope this is something that can be corrected to allow the logging to be used more effectively for long term experiments.

     

    Secondly, in order to start the logger I have to have a USB stick installed. What if I don't have a USB stick available? The logger then becomes unusable. This is a network connected device so why not have the ability to save to a network location or even use some remote logging server which the unit can connect to easily to store the logs?

     

    Finally, I have noticed some waveform glitches on the display, regularly part of a trace will disappear for a second and then reappear. It's quite distracting having a flashing trace on the log view.

     

    As a quick follow up to the earlier comment about the power trace not making it into the CSV file, I have not discovered the reason for this yet. If it's there as a trace I can enable for logging it should be in the file. Yes, I can create it from the V and I for the trace but if I have a long data set adding that column to my spreadsheet is fiddly and involves lots of scrolling. This should also be fixed. While I still think the logging is a great feature and very welcome, it isn't bug free and does have some limitations so for now I'm going to amend my review score down a point for the bugs and the newly discovered logging limitations. If a new firmware is released that resolves these issues I will report back and amend the score accordingly.

     

    5.6. Using the Rear Outputs for 4-Wire Connections

    5.6.1. Why 4-Wire Connections?

     

    In an ideal world your wires which connect your power supply to your load would have zero or negligible impedance, but in reality there is a small impedance present, usually in the order of a few tens or hundreds of mOhm for typical bench hook up cables. For low current applications this small impedance makes almost no difference and the voltage measured directly at the load is almost identical to the voltage measured at the output of the power supply. As the current increases, using Ohms law (V = I/R) we can soon see how higher currents could lead to a significant drop in the voltage between the source and the load. In order to compensate, if the load was constant you could just raise the source voltage until the load voltage was at the desired level, but this is not a suitable approach as the load could change and you may end up with an over-voltage condition on the input to your load.

     

    Usually the power supply will feed back the voltage it is using to control the output of the supply from as close to the output as possible. This compensates for any losses in the output stage of the supply but does nothing to help the case where there are additional external losses. This is where 4-Wire connections come into play. An additional pair of wires connect from the power supply to the load and these are used as dedicated sense pairs to measure the voltage directly at the load. These sense wires carry none of the output current so there is no current related voltage drop, it's an accurate measurement of the voltage directly at the load. This measurement is used instead (or as well as) of the internal voltage measurement to enable to supply to automatically deliver the desired voltage directly at the load.

     

    5.6.2. The 4-Wire Test Setup and Results

     

    Keysight E36313A ConfigurationKeysight E36313A Results

     

    With the output set to 2-Wire mode with a 5V output and 6.66A load (as reported on the TTi LD300 DC Electronic Load) the voltage measured on my DMM is 4.428V. This means that in my cabling between the power supply output and the load there is almost over 0.57V dropped. This is plenty to cause issues for test devices which require 5V +/-10%.

     

    Update 20/12/2017: Having thoroughly investigated my strange readings I have discovered there was an issue with the measurements on my DMM and the voltages I was reading were not correct. Switching my TTi LD300 into meter display so it shows the voltage it sees on it's input and the current it is sinking, with a 10A load and in 2-Wire mode, the voltage at the input to the load is just 4.05V. If I vary the load the measured voltage varies as expected.

     

    So it turns out the optimal test current for my particular the 4-Wire setup to be able to correct back up to exactly 5.000V as measure on my DMM is 6.66A. Above this and the PSU doesn't have enough headroom to be able to recover the voltage drop. But up to this, in my setup and the supply perfectly corrected back to exactly 5.000V.

     

    Now to a slight issue, I then dropped the load to only 1A and the supply now incorrectly gives 5.168V measured on my DMM at the load. I need to do some more investigation on this as this is clearly not correct. I need to double check all my test setup and do some more analysis and see if I can fathom what is happening here.

     

    Update 20/12/2017: As stated above, the readings from my DMM were incorrect. On the TTi LD300 meter view, with a 10A load and 4-Wire connections enabled the reading is now exactly 5.00V. This measurement now remains constant regardless of the load.

     

    6. Road Test Summary and Conclusions

     

    All in all I think this unit is very good. It packs in a lot of features and provides a really nice to use interface to access the features. My unit came with the USB/LAN and GPIB interfaces to connect with other systems easily and this opens up a whole lot of additional possibilities. I initially knocked off a couple of points in the "Performed to Expectations" section, one because of the mains hum from my unit, and one for the little niggles I discovered in the UI. Keysight will hopefully provide a firmware update to fix the UI issues at some point. With their response regarding the hum issue and their prompt action to replace my unit I am adding back the point I deducted for this as it appears to be specific to my unit and they are in the process of resolving the issue for me.

     

    In comparison to the TTi units I already have you can see that it performs well. To be fair to TTi, all three unit's are very good but I think the Keysight E36313A out performs the TTi units in a few areas and the user interface and feature set is a deciding factor for me. The output can be seen from the various load tests and FFT's to be a little cleaner and some higher frequency artefacts which are present on the output of the TTi supplies are either not present or much reduced on the Keysight E363213A in comparison.

     

    One thing I can't do with my TTi supplies is easily log data and this is a very valuable tool which the Keysight E36313A makes really easy to use. I can easily log the Voltage, Current, and Power from each of the three channels and then use this data to present these values in a useful and meaningful way or to process the data to extract hidden information such as tracking the overall power consumption of a low power device with many different power states over time.

     

    This concludes my Road Test Review of the Keysight E36313A. Thank you all for reading this review and thanks again to Element 14 and Keysight for choosing me for this Road Test. It has been an honour to receive this item and a pleasure to go through and do a comprehensive review of this excellent piece of equipment. It's now 100% my go to power supply in my lab.

     

    20/12/2017: I've updated the section above regarding the 4-Wire connections to correct my test results and report that is it in fact working well and exactly as expected.

    21/12/2017: I've updated the section above regarding the logging capabilities to add my additional findings and limitations.


Comments

Also Enrolling
Enrollment Closes: Nov 26 
Enroll
Enrollment Closes: Oct 24 
Enroll
Enrollment Closes: Oct 26 
Enroll
Enrollment Closes: Nov 12 
Enroll
Enrollment Closes: Nov 16 
Enroll
Enrollment Closes: Oct 24 
Enroll
Enrollment Closes: Nov 1 
Enroll
Enrollment Closes: Nov 5 
Enroll
Enrollment Closes: Oct 24 
Enroll